A new method was developed for the analysis of opioids in urine samples. The following data was obtained via the calibration curve method for opioid

A new method was developed for the analysis of opioids in urine samples. The following data
was obtained via the calibration curve method for opioid A.
)
Signal intensity +- standard deviation (a. u.)*
0
20+-6.6
5
30+-4.5
10
40+-2
20
65+-5
40
105+-10.5
60
150+-22.5
80
Calculate the limit of detection (LOD) and the limit of quantitation (LOQ) of the method.
a) LOD=9.1ppm;LOQ=30.4ppm
b) LOD=10.6ppb;LOQ=35.2ppb
c) LOD=9.1ppb;LOQ=30.4ppb
d) LOD=10.6ppm;LOQ=35.2ppm
e) None of the above.
What is the centroid of the calibration curve?
a)(20,65)
b)(40,105)
c)(30.7,81.4)
d)(22.5,68.3)
e) None of the above.
The number of blank determinations was 16. All the other signals are the average of 3
determinations each.analytical Chemistry.
image text in transcribed

Step by Step Solution

There are 3 Steps involved in it

Step: 1

blur-text-image

Get Instant Access to Expert-Tailored Solutions

See step-by-step solutions with expert insights and AI powered tools for academic success

Step: 2

blur-text-image_2

Step: 3

blur-text-image_3

Ace Your Homework with AI

Get the answers you need in no time with our AI-driven, step-by-step assistance

Get Started

Recommended Textbook for

Introduction to Chemical Engineering Thermodynamics

Authors: J. M. Smith, H. C. Van Ness, M. M. Abbott

7th edition

71247084, 978-0071247085

More Books

Students also viewed these Chemical Engineering questions

Question

Describe the biopsychosocial approach to individual development.

Answered: 1 week ago