You want to prepare static-test-atmospheres in Tedlar bags in order to calibrate a photoionization detector (PID). You are going to use the PID to

image text in transcribed 

You want to prepare static-test-atmospheres in Tedlar" bags in order to calibrate a photoionization detector (PID). You are going to use the PID to evaluate worker exposure to perchloroethylene in a dry-cleaning plant The current ACGIH TLV for perchloroethylene (CAS: 127-18-4) is 25 ppm You want to prepare three different concentrations bracketing the TLV in order to calibrate. Your target concentrations are %, 1, and 2x the occupational exposure limit (OEL). You will use a dry-gas meter to introduce five liters of clean air into each of the three bags, and a microliter syringe to transfer the perchloroethylene (o 1.62 g/mL). Calculate the required volume of liquid perchloroethylene that you will add to each bag. Bag #1: Concentration= Bag #2: Concentration= Bag #3: Concentration= Quantity of perchlorocethylene (uL)_ Quantity of perchloroethylene (ul)_ Quantity of perchloroethylene (uL)_

Step by Step Solution

3.33 Rating (150 Votes )

There are 3 Steps involved in it

Step: 1

TLV for perchloroethylene is 25 ppm Bag 1 concentration 1x TLV 25 ppm Bag 2 concentrat... blur-text-image

Get Instant Access to Expert-Tailored Solutions

See step-by-step solutions with expert insights and AI powered tools for academic success

Step: 2

blur-text-image

Step: 3

blur-text-image

Ace Your Homework with AI

Get the answers you need in no time with our AI-driven, step-by-step assistance

Get Started

Recommended Textbook for

Investments

Authors: Zvi Bodie, Alex Kane, Alan Marcus, Stylianos Perrakis, Peter

8th Canadian Edition

007133887X, 978-0071338875

More Books

Students also viewed these Finance questions