sequential test data 10 K = 0.00011996 V/Pa 8 6 A pressure transducer has the following specifications: Input range: 0 to 50 kPa Excitation: +15V dc Output range: 0 to 10 V de Thermal sensitivity error = 0.02% PC of reading Thermal zero drift = 0.05% !C FSO Repeatability = 0.05% FSO Resolution = +0.1V de Temperature range: 0 to 50C The pressure transducer will be used to measure pressures varying from 0 to 30 kPa and at temperatures varying from 15 to 25 C. a. find the linearity error from the sequential data given to the right b. find the hysteresis error from the sequential data given to the right c. find the magnitude of each individual error d. calculate the overall instrument uncertainty voltage, V 2 0 -2 0 10 40 50 20 30 pressure, kPa Red line-linear fit to increasing data Blue line-data when pressure is being increased Green line - data when pressure is being decreased sequential test data 10 K = 0.00011996 V/Pa 8 6 A pressure transducer has the following specifications: Input range: 0 to 50 kPa Excitation: +15V dc Output range: 0 to 10 V de Thermal sensitivity error = 0.02% PC of reading Thermal zero drift = 0.05% !C FSO Repeatability = 0.05% FSO Resolution = +0.1V de Temperature range: 0 to 50C The pressure transducer will be used to measure pressures varying from 0 to 30 kPa and at temperatures varying from 15 to 25 C. a. find the linearity error from the sequential data given to the right b. find the hysteresis error from the sequential data given to the right c. find the magnitude of each individual error d. calculate the overall instrument uncertainty voltage, V 2 0 -2 0 10 40 50 20 30 pressure, kPa Red line-linear fit to increasing data Blue line-data when pressure is being increased Green line - data when pressure is being decreased